1. Reagents
 Standards:acrylamide, for electrophoresis (Kanto Chemical Co., Inc., Tokyo, Japan), purity 99.9% ; internal standard, acrylamide-1-13C (AA-1-13C) (CDN ISOTOPES Co., Quebec, Canada), 99 atom% 13C.
 Calibration standard solution:Aliquots of AA standard solutions (equivalent to 0, 50, 100, 500, 1,000, 2,000 and 5,000 ng of AA) were mixed with 20 μL of internal standard solution (100 μg/mL AA-1-13C) and made up to 10 mL with water.
 Bond Elut C18 (500 mg/3 mL), Bond Elut Jr-PSA, (500 mg) and Bond Elut ACCUCAT (600 mg/3 mL) : Bond Elut C18 was conditioned with 2 mL of methanol and successively with 4 mL of water before use. The other columns were conditioned only with 4 mL of water before use.
 Acetone: high quality for pesticide residue analysis. Dichloromethane, methanol and water: HPLC grade.

2. Apparatus
 Four columns were used and switching valves were inserted between column-1 and -2, and column -2 and -3.
 Liquid chromatograph/mass spectrometer:model 2690 separation module liquid chromatograph (Waters, Co.) coupled with a model Platform LCZ mass selective detector (Micromass UK Ltd.)
 Degasser: ERC-3215α (ERC Inc., Kawaguchi, Japan)
 Switching valve: LabPRO (Rheodyne) and EV700-100 (GL Sciences)
 Pump: LC-10AD and LC-10ADvp ( Shimadzu Co.)
 Homogenizer:Polytron (KINEMATICA AG, Littau, Switzerland)
 Concentrator: Turbo Vap 500 (Zymark Co)

3. LC/MS conditions
 Column: column-1, Hypercarb(50×2.1 mm i.d., particle size 5 μm, Thermo Hypersil-Keystone); column-2, Shodex MSpak GF-310 4B (50×4.6 mm i.d., particle size 6 μm, Shoko. Co. LTD.); column-3, Atlantis dC18 (150×2.1 mm i.d., particle size 5 μm, Waters Co.); column-4, Develosil RPAQUEOUS-AR-3 (35×2.0 mm i.d., particle size 3 μm, Nomura Chemical Co.)
 Column oven temp.: 40°C for the column-1, -2 and -3, and room temperature for the column-4
 Mobile phase composition: 0.1% acetic acid-methanol (95:5)
 Flow rate; 0.25 ml/min
 Sampling time from column-1 to column-2: 1.75〜2.05 min
 Sampling time from column-2 to column-3 : 4.66〜5.15 min
 AA eluting time from column-4: 8.5 min
 Source block temp.: 120°C, Desolvation temp.: 350°C, Gas: nitrogen, flow rate, ca.300 L/hr, Capillary voltage: 3.50 eV, Ionization mode: positive ion electron spray,  Measuring mode: selected ion recording(SIR), Monitoring ion and corn voltage : m/z 72 (AA) and m/z 73 (AA-1-13C ) at 24 V of corn voltage and m/z 55 (AA) and m/z 56 (AA-1-13C ) at 40 V of corn voltage, Injection volume: 10 mL

4. Sample preparation
 Sample (5.0 g) was weighed into a 250-mL glass tube and 20〜25 mL of water and 100 μL of internal standard solution (100 μg/mL AA-1-13C solution) were added to the sample. The mixture was allowed to stand for 30 min with occasional shaking. Acetone (50 mL) was added and the mixture was homogenized for 1 min with a homogenizer. After filtrating the homogenate through a glass filter (GA-100), the residue on the filter was washed with 30 mL of acetone. The filtrate was concentrated to less than 15 mL by a concentrator at 35℃. The extract was transferred in a 50-mL centrifugal tube, washed twice with each 20 mL of dichloromethane by shaking for 1 min and centrifuged at 2,700 rpm for 5 min. The extract was made up to 15 mL with water.
 An aliquot of the extract (3 mL, equivalent to 1.0 g of sample) was passed through a C18 cartridge column with following washing with 2 mL of water. All the eluate was collected and made up to 5 mL with water. An aliquot of the eluate (2 mL, equivalent to 0.4 g of sample) was passed through an ACCUCAT column connected to PSA column with following washing with 2 mL of water. All the eluate was collected and made up to 4 mL with water to prepare the test solution.

5. Determination and confirmation of AA
 AA was determined by LC/MS(SIR). The quantification was performed by the comparison of the peak area ratios of m/z 72 (AA) and m/z 73 (AA-1-13C) for the test solution with those for the calibration standard. The peak area ratio of m/z 55 (AA) and m/z 56 (AA-1-13C) was used for the qualification. The recovery of internal standard was calculated from the peak areas of AA-1-13C in the test solution and the calibration standard.
 The calibration standard (AA=0, AA-1-13C=200 ng/mL) contained a small quantity of AA as an impurity of AA-1-13C. Therefore, the limit of detection (LOD) was defined as 3 times the SD of the AA peak area of calibration standard (AA=0, AA-1-13C=200 ng/mL). Similarly, the limit of quantification (LOQ) was defined as 10 times the SD.